The purpose of this study is to achieve the quantitative detection of recombinant type III collagen (rh-COL-III) in dressings with complex matrix. First of all, the marker peptide (GEAGIPGVPGAK) of rhCOL-III was identified with HPLC-MS/MS. Then, a qualitative and quantitative method based on marker peptides was established and validated. In order to obtain higher sensitivity, a pretreatment method of liquid, gel, and ointment dressings was optimized. The reference material for quantification was combined using rhCOL-III and blank matrix of each dressing. The results indicated that the relative standard deviation (RSD) of the quantitative method was 2.77%, and the RSD of intraday and interday precision was 2.76% and 2.31%, respectively. The spiking recovery rate was between 80% and 90%. The optimal pretreatment method was Tris-HCl solvent replacement. The optimal trypsin concentration for the dressing solution was 20 μg in 500 μL. The method of preparing standard substances with a blank matrix can effectively eliminate the influence of the matrix effect on the quantitative results. The average spiking recovery rates of 50 μg/mL, 100 μg/mL, and 200 μg/mL in three different dressings ranged from 80% to 120%. The quantitative detection of limit (LOD) of rhCOL-III was 1 ng/mL, 2 ng/g, and 1 ng/g in liquid, ointment, and gel dressings.