The purpose of this study was to implement a state-of-the-art convolutional neural network used to synthesize artificial T1-weighted (T1w) full-dose images from corresponding noncontrast and low-dose images (using various settings of input sequences) and test its performance on a patient population acquired prospectively. In this monocentric, institutional review board-approved study, a total of 138 participants were included who received an adapted imaging protocol with acquisition of a T1w low dose after administration of 10% of the standard dose and acquisition of a T1w full dose after administration of the remaining 90% of the standard dose of a gadolinium-containing contrast agent. A total of 83 participants formed the training sample (51.7 ± 16.5 years, 36 women), 25 the validation sample (55.3 ± 16.4 years, 11 women), and 30 the test sample (55.0 ± 15.0 years, 9 women). Four input settings were differentiated: only the T1w noncontrast and T1w low-dose images (standard setting), only the T1w noncontrast and T1w low-dose images with a prolonged postinjection time of 5 minutes (5-minute setting), multiple noncontrast sequences (T1w, T2w, diffusion) and the T1w low-dose images (extended setting), and only noncontrast sequences (T1w, T2w, diffusion) were used (zero-dose setting). For each setting, a deep neural network was trained to synthesize artificial T1w full-dose images, which were assessed on the test sample using an objective evaluation based on quantitative metrics and a subjective evaluation through a reader-based study. Three readers scored the overall image quality, the interchangeability in regard to the clinical conclusion compared with the true T1w full-dose sequence, the contrast enhancement of lesions, and their conformity to the respective references in the true T1w full dose. Quantitative analysis of the artificial T1w full-dose images of the standard setting provided a peak signal-to-noise ratio of 33.39 ± 0.62 (corresponding to an average improvement of the low-dose sequences of 5.2 dB) and a structural similarity index measure of 0.938 ± 0.005. In the 4-fold cross-validation, the extended setting yielded similar performance to the standard setting in terms of peak signal-to-noise ratio ( P = 0.20), but a slight improvement in structural similarity index measure ( P < 0.0001). For all settings, the reader study found comparable overall image quality between the original and artificial T1w full-dose images. The proportion of scans scored as fully or mostly interchangeable was 55%, 58%, 43%, and 3% and the average counts of false positives per case were 0.42 ± 0.83, 0.34 ± 0.71, 0.82 ± 1.15, and 2.00 ± 1.07 for the standard, 5-minute, extended, and zero-dose setting, respectively. Using a 5-point Likert scale (0 to 4, 0 being the worst), all settings of synthesized full-dose images showed significantly poorer contrast enhancement of lesions compared with the original full-dose sequence (difference of average degree of contrast enhancement-standard: -0.97 ± 0.83, P = <0.001; 5-minute: -0.93 ± 0.91, P = <0.001; extended: -0.96 ± 0.97, P = <0.001; zero-dose: -2.39 ± 1.14, P = <0.001). The average scores of conformity of the lesions compared with the original full-dose sequence were 2.25 ± 1.21, 2.22 ± 1.27, 2.24 ± 1.25, and 0.73 ± 0.93 for the standard, 5-minute, extended, and zero-dose setting, respectively. The tested deep learning algorithm for synthesis of artificial T1w full-dose sequences based on images after administration of only 10% of the standard dose of a gadolinium-based contrast agent showed very good quantitative performance. Despite good image quality in all settings, both false-negative and false-positive signals resulted in significantly limited interchangeability of the synthesized sequences with the original full-dose sequences.