PurposeA new mobile cone beam computed tomography (CBCT) imaging ring (IRm, Elekta, v2.10.6, Veenendaal, Netherlands) has recently been proposed for brachytherapy to improve procedure efficiency. We describe the commissioning process and end-to-end tests for GYN HDR brachytherapy employing IRm CBCT imaging. Materials and MethodsCommissioning included imaging isocenter test, 3D image quality, 2D imaging quality, image dose, and tube characteristics. CIRS HDR GYN phantom and Venezia CT/MR gynecological applicator were used to perform end-to-end (E2E) tests and optimize workflow. Venezia applicator and four interstitial needles were inserted into the phantom and IRm CBCT images were acquired. Phantom and applicator were scanned with CT scanner (Siemens SOMATOM go.Open Pro) using department's pelvis imaging protocol. MR imaging was performed using 0.35T MR Linac TRUFI pulse sequence. CBCT images were registered to CT and MR using rigid registration to assess image quality and applicator geometry fidelity. ResultsAll physics tests passed within acceptance tolerances. Registration of CBCT images to MR and CT scans was acceptable for applicator placement. Applicator registration of CBCT images to CT demonstrated excellent agreement of most distal source dwell position (<1 mm). Slice thickness was also measured to be 1.25 mm, within 0.5 mm of its nominal value. ConclusionBased on E2E and commissioning results, IRm is an appropriate tool for brachytherapy treatment planning. This study demonstrated good image quality in GYN phantom and Venezia applicator using the IRm. Distal source dwell position agreement between CBCT and CT was acceptable for clinical use.