Recently, the Bonebridge 602 implant was introduced. Its smaller dimensions facilitate implantation even in surgically demanding cases. However, in extreme anatomical conditions, implant lifts are still required. We intended to report on the medical and audiological outcomes of all patients implanted with the implant with a secondary focus on the safety, efficacy, and feasibility of the use of 1-mmlifts. Retrospective cohort study. Tertiary academic center. We retrospectively analyzed all patients implanted with the Bonebridge 602. Patient demographics and surgically-specific data were collected from the medical charts. Furthermore, audiological results were assessed. Outcomes were compared between implantations with versus without lifts. Twenty-one devices were implanted during the study period. Satisfactory audiological results were observed, while no intra- or postoperative adverse events occurred. The majority of patients were daily users at the end of the observation period (n = 20/21, 95.2%), and one patient (n = 1/21, 4.8%) was a nonuser after 6 months of use because of subjective dissatisfaction. Due to anatomical considerations, 1-mm lifts were used in 4 implantations (19.0%). The application of lifts did not result in prolonged surgical times, complications, or shorter time of use, nor did it negatively affect audiometric results. Implantations with the new and smaller Bonebridge were associated with gratifying medical and audiological outcomes. Still, in extreme anatomical conditions, 1-mm lifts are necessary. Nonetheless, surgical placement with the help of lifts seems safe, effective, and feasible, and is a viable option in cases with challenging anatomies due to previous surgeries or anomalies.