BackgroundTranscatheter aortic valve implantation is unfeasible for 10–15% of patients using the conventional transfemoral approach. Other alternative approaches, such as the subclavian approach, have emerged, with no clear recommendation indicating the superiority of one technique over another. AimTo compare the 1-month mortality and postprocedural outcomes of patients undergoing transcatheter aortic valve implantation using a self-expandable valve via transfemoral and subclavian access. MethodsThis was a retrospective single-centre study including 1496 patients who underwent transcatheter aortic valve implantation between January 2016 and December 2020 at Clermont-Ferrand University Hospital, France. Propensity score matching was used to compare transfemoral and subclavian access. ResultsAfter building two propensity score-matched groups of 221 patients each with either access route (total n=442), baseline characteristics were similar. The procedure duration was significantly longer in the subclavian access group (53 [45–64] versus 60 [51–72] minutes; P<0.001), but with a lower amount of contrast agent (138 [118–165] versus 123 [105–150] mL; P<0.001), fluoroscopy time (11.2 [9–14] versus 9.9 [7–12] minutes; P<0.001) and radiation dose (397 [264–620] versus 321 [217–485] mGy; P<0.001). No significant difference was observed concerning 1-month mortality (odds ratio 1.62, 95% confidence interval 0.52–5.03; P=0.39) or periprocedural complications. Follow-up at 1 year confirmed no difference in longer-term mortality (hazard ratio 0.78, 95% confidence interval 0.52–5.03; P=0.43). ConclusionsThe subclavian approach provides similar results to the transfemoral approach in terms of mortality, efficacy and safety; it is a reasonable and effective alternative when the reference transfemoral approach is impossible or seems complex.