To compare the accuracy of modern intraocular lens (IOL) power calculation formulas with that of older formulas, such as SRK/T and Hoffer Q, in pediatric cataract surgery. This retrospective study included 100 eyes of 100 children who underwent routine cataract surgery with primary IOL implantation in a bag. This study used four IOLMaster 700 integrated formulas: SRK/T, Hoffer Q, Haigis, and Barrett Universal II (BUII). In addition, the following formulas were used: EVO 2.0, Hill RBF 3.0, Hoffer QST, Kane, and PEARL DGS, which are available online. There was a statistically significant difference between SRK/T and most other formulas, except for Hoffer Q, Hoffer QST, and BUII (p < 0.05). SRK/T yielded the lowest median absolute error (MedAE) of 0.63 D. This was followed by the BUII (0.66 D), Hoffer Q, and Hoffer QST (0.68 D). SRK/T also yielded the highest percentage of cases within ± 0.50 D (43% of the cases). For patients aged 2 to 5 years, SRK/T formula yielded statistically significantly better results than all other included formulas (p < 0.05) with MedAE = 0.44 D, 58.33% and 87.50% of the cases were within ± 0.50 D and ± 1.0 D of intended refraction, respectively. The SRK/T formula showed the best IOL power calculation results in pediatric cataract surgery, followed by BUII, Hoffer Q, and Hoffer QST. In children aged 2-5 years, the SRK/T formula outperformed all other formulas, followed by the BUII and Hoffer QST formulas. In children older than 5 years, there was no statistically significant difference between the different formulas (p > 0.05); Hoffer Q and SRK/T showed slightly better MedAE in this age group (5-10 years).