PurposeTo investigate the outflow characteristics of silicone tubes with intraluminal stents used in membrane-tube (MT) type glaucoma shunt devices.MethodsThe silicone tubes used in MicroMT (internal diameter of 100 µm with a 7-0 nylon intraluminal stent) and Finetube MT (internal diameter of 200 µm with a 5-0 nylon intraluminal stent) were connected to a syringe-pump that delivered a continuous flow of distilled water at flow rates of 2, 5, 10, and 25 µL/min. The pressures and resistances of tubes were measured at a steady flow rate with full-length, half-length, and absence of intraluminal stents.ResultsThe mean outflow resistance of the two types of tubes ranged from 3.0 ± 1.9 to 3.8 ± 1.7 mmHg/µL/min with a full-length intraluminal stent, 1.8 ± 1.1 to 2.2 ± 1.1 mmHg/µL/min with a half-length intraluminal stent, and 0.1 ± 0.0 to 0.2 ± 0.0 mmHg/µL/min without an intraluminal stent. Theoretically, for a physiologic state with a flow rate of 2 µL/min and episcleral venous pressure of 6 mmHg, the mean pressures of tubes were expected to be 13.2 ± 3.0, 10.5 ± 2.4, and 6.4 ± 0.2 mmHg in MicroMT with full-length, half-length, and absence of intraluminal stents, respectively, and 12.5 ± 3.9, 9.6 ± 2.4, and 6.2 ± 0.2 mmHg in Finetube MT with full-length, half-length, and absence of intraluminal stents, respectively. The pressure variance also decreased with intraluminal stent retraction (p < 0.01).ConclusionsThe small diameter tubes of 100 and 200 µm internal diameters, with 7-0 and 5-0 nylon intraluminal stents, respectively, used in the MT-type glaucoma shunt device showed safe and effective outflow characteristics.