A technical comparison is described between two SPECT systems, one dedicated to cardiovascular studies and one general-purpose, to evaluate the advantages and disadvantages of their use in an organ-specific clinical setting. The comparison was made between a dedicated cardiac SPECT (Alcyone, Discovery NM 530c, GE Healthcare) scanner and a general-purpose one (Discovery NM/CT 670 CZT, GE Healthcare). The two scanners differ in terms of hardware, mainly in the arrangement of the detectors and collimators, which are the main components of SPECT. A standard NEMA phantom was used to characterize the energy resolution, spatial resolution, and sensitivity for the two systems. Then, using a custom-made cardiac phantom, more specific indices were computed to evaluate the quality of cardiac images, such as signal-to-background noise ratio (SBNR), tissue-background contrast-to-noise ratio (TBCNR), and uniformity. Finally, the same indices were computed for clinical images acquired with the two systems from 11 subjects. Alcyone showed superior performance for dedicated cardiac imaging; however, its excellent qualities are aimed only at the study of the heart and only at patients with a low body-mass index, unlike Discovery NM/CT 670 CZT, which can be used for every anatomic district area and for every type of patient.