To compare the image quality of short-tau inversion recovery (STIR) and the STIR-slice encoding for metal artifact correction (SEMAC) sequence for postsurgery spine magnetic resonance imaging (MRI). Twenty-nine patients with metallic spinal implants who underwent spinal 1.5 T MRI with STIR and STIR-SEMAC sequences between July 2016 and November 2020 were retrospectively enrolled. Qualitative assessments were performed using 5-point scales; higher scores indicated better image quality. For screw metal artifact analysis, scores were obtained for artifacts on vertebral bodies and neural foramina, screw artifact widths, and bone marrow signal intensities. For patient-based analysis, scores were obtained for imaging quality and fat suppression quality, signal intensity, and cerebrospinal fluid noise. A paired t test was performed for statistical analyses. We analyzed 163 screws in 29 patients. In the screw metal artifact analysis, the vertebral body and neural foramen scores were significantly higher for the STIR-SEMAC images than for the STIR (all P < 0.001). The artifact width in the STIR-SEMAC images (9.8 ± 3.4 mm) was significantly smaller than that in the STIR images (16.0 ± 4.7 mm, P < 0.001). In patient-based analysis, the fat suppression and imaging quality scores were significantly higher for the STIR-SEMAC images than for the STIR images (all P < 0.001). The cerebrospinal fluid signal intensity, noise, and signal-to-noise ratios were significantly higher for the STIR images (all P < 0.005). Short-tau inversion recovery-SEMAC sequences provide good metallic artifact reduction and fat suppression for postsurgery spine 1.5 T MRI.