To determine the feasibility of simultaneous multi-slice (SMS) real-time MRI (RT-MRI) at 0.55T for the evaluation of cardiac function. Cardiac CINE MRI is routinely used to evaluate left-ventricular (LV) function. The standard is sequential multi-slice balanced SSFP (bSSFP) over a stack of short-axis slices using electrocardiogram (ECG) gating and breath-holds. SMS has been used in CINE imaging to reduce the number of breath-holds by a factor of 2-4 at 1.5T, 3T, and recently at 0.55T. This work aims to determine if SMS is similarly effective in the RT-MRI evaluation of cardiac function. We used an SMS bSSFP pulse sequence with golden-angle spirals at 0.55T with an SMS factor of three. We cover the LV with three acquisitions for SMS, and nine for single-band (SB). Imaging was performed on 9 healthy volunteers and 1 patient with myocardial fibrosis and sternal wires. A spatio-temporal constrained reconstruction is used, with regularization parameters selected by a board-certified cardiologist. Images were quantitatively analyzed with a normalized contrast and an Edge Sharpness (ES) score. There was a statistically significant 2-fold difference in contrast between SMS and SB and no significant difference in ES score. The contrast for SMS and SB were 13.38/29.05 at mid-diastole and 10.79/22.26 at end-systole; the ES scores for SMS and SB were 1.77/1.83 at mid-diastole and 1.50/1.72 at end-systole. SMS cardiac RT-MRI at 0.55T is feasible and provides sufficient blood-myocardium contrast to evaluate LV function in three slices simultaneously without any gating or periodic motion assumptions.