PurposeThe cortical uptake of tau positron emission tomography (PET) tracers corresponds to the Braak stage and reflects the distribution and progression of tau neurofibrillary tangles. The present study aimed to develop and validate the basic performance of a novel tau PET phantom, as well as to establish standard test procedures and analytical methods. MethodsThe tau PET phantom consisted of a brain simulation section simulated medial temporal lobe region and resolution and uniformity sections. The brain simulation section and hot rods and uniformity section contained 4 and 2 kBq/mL of 18F, respectively and images were acquired three times for 20 min with a PET/CT scanner. The resolution section was visually assessed with two sets of hot and cold rods. Recovery coefficients (RCs) as a quantitative value and coefficient of variation (CV) as image noise were determined based on the brain simulation and the uniformity section, respectively. ResultsPreparation of activity in the phantom was repeatable among three measurements. The quality of images in the brain simulation and uniformity section with the rods was good. The 5- or 6-mm rods were detected separately. The mean RCs calculated based on the VOI template were between 0.75 and 0.83. The CV at the center slice of uniformity section was 5.54%. ConclusionsWe developed a novel tau PET phantom to assess quantitative value, image noise, and detectability and resolution from brain simulation section, uniformity section, and rods, respectively. This phantom will contribute to the standardization and harmonization of tau PET imaging.