Intraoperative imaging is becoming increasingly common in repair of facial fractures. Many institutions do not have access to intraoperative advanced 3D imaging but have the capability of obtaining plain radiographs intraoperatively. At institutions where advanced 3D imaging is available, scout radiographs are usually obtained prior to a complete scan. These scout images can provide some information about the placement of radiopaque implants before a complete scan is performed. The aim of this study is to examine the correct anatomic positioning of an orbital floor implant using lateral plain radiographs. Titanium orbital fan implants were molded and secured to orbital floor of 14 adult dry skulls (7 males and 7 females). Lateral radiographs were obtained for both the left and the right orbits individually. The antero-posterior angle of inclination that the implant makes relative to the Frankfort horizontal plane was measured, and results were compared in the male versus female radiographs. The mean angle that the implant made with the Frankfort horizontal plane was 20.1±2.4° in the male orbits (95% CI 18.8-21.5°) and 22.6 ± 2.0° in the female orbits (95% CI, 21.4-23.7°). We found no statistically significant differences between the male and female angles (P-value 0.62). For the combined specimens (i.e., 28 sides in 14 skulls), the overall mean angle was 21.4 ± 2.5° (95% CI, 20.4-22.3°). Based on these findings, the angle that the implant makes with the Frankfort horizontal plane on lateral cross-table plain radiographs can be used intraoperatively to assist the surgeon in confirming the appropriate placement of the implant.