The purpose of this study is to evaluate the RF field responses of partial-ring RF-shielded oval-shaped positron emission tomography (PET) inserts that are used in combination with an MRI body RF coil. Partial-ring PET insert is particularly suitable for interventional investigation (e.g., trimodal PET/MRI/ultrasound imaging) and intraoperative (e.g., robotic surgery) PET/MRI studies. In this study, we used electrically floating Faraday RF shield cages to construct different partial-ring configurations of oval and cylindrical PET inserts and performed experiments on the RF field, spin echo and gradient echo images for a homogeneous phantom in a 3T clinical MRI system. For each geometry, partial-ring configurations were studied by removing an opposing pair or a single shield cage from different positions of the PET ring. Compared to the MRI-only case, reduction in mean RF homogeneity, flip angle, and SNR for the detector opening in the first and third quadrants was approximately 13%, 15%, and 43%, respectively, whereas the values were 8%, 23%, and 48%, respectively, for the detector openings in the second and fourth quadrants. The RF field distribution also varied for different partial-ring configurations. It can be concluded that the field penetration was high for the detector openings in the first and third quadrants of both the inserts.