Reverse transcription-digital PCR (RT-dPCR) is attracting attention as a method that enables SI-traceable RNA quantification without calibration, but its accuracy and bias have not been thoroughly studied. In this study, the accurate quantification of RNA by the RT-dPCR method was investigated using NMIJ CRM 6204-b, an RNA certified reference material whose certified value was assigned by orthogonal chemical measurement methods. Moreover, a two-step RT-dPCR method was adopted to examine in detail the conditions for the RT reaction process, which was expected to be the major uncertainty component in the RT-dPCR measurement. Optimization experiments revealed that the type of reverse transcriptase, the concentration of template RNA, and the type and concentration of primers in the RT reaction affected the value quantifiedby RT-dPCR. Under the optimal conditions, the value quantifiedby RT-dPCR, 76.4ng/μL ± 6.7ng/μL (the quantified value ± expanded uncertainty (k = 2)), was consistent with the certified value, 68.2ng/μL ± 5.8ng/μL, of NMIJ CRM 6204-b RNA 1000-A within the expanded uncertainty. From the results of the uncertainty evaluation, the relative combined uncertainty of the RT-dPCR method was 4.42%, and the major uncertainty components in the RT-dPCR method were the preparation of RT solution (3.68%), the inter-day difference (1.80%), and the RT reaction (1.30%). Together, the results suggested that the contribution of the RT reaction process to the total uncertainty was greater than that of the dPCR process.