This study evaluates intracochlear electrocochleography (ECochG) for real-time monitoring during cochlear implantation. One aim tested whether adjusting the recording electrode site would help differentiate between atraumatic and traumatic ECochG amplitude decrements. A second aim assessed whether associations between ECochG amplitude decrements and post-operative hearing loss were weaker when considering hearing sensitivity at the ECochG stimulus frequency compared to a broader frequency range. Eleven adult cochlear implant recipients who were candidates for electro-acoustic stimulation participated. Single-frequency (500-Hz) ECochG was performed during cochlear implantation; the amplitude of the first harmonic of the difference waveform was considered. Post-operative hearing preservation at 500 Hz ranged from 0 to 94%. The expected relationship between ECochG amplitude decrements and hearing preservation was observed, though the trend was not statistically significant, and predictions were grossly inaccurate for two participants. Associations did not improve when considering alternative recording sites or hearing sensitivity two octaves above the ECochG stimulus frequency. Intracochlear location of a moving recording electrode is a known confound to real-time interpretation of ECochG amplitude fluctuations, which was illustrated by the strength of the correlation with ECochG amplitude decrements. Multiple factors contribute to ECochG amplitude patterns and to hearing preservation; these results highlight the confounding influence of intracochlear recording electrode location on the ECochG.