ABSTRACT With the continuous development of biomedicine, the long-term low-temperature storage of biological samples has become essential. In order to meet the demand for long-term storage of liquid nitrogen to keep samples safe, a study on the evaporated gas recycling storage of liquid nitrogen tanks will be carried out. This research utilizes new approach to accurately calculate the heat leakage of the liquid nitrogen tank. The experimental results show that the theoretical heat leakage is 6.16 W, and the actual heat leakage is 7.07 W. By incorporating the insulating plug, the actual heat leakage was reduced by 61.2%, decreasing to 2.74 W, and the calculated heat leakage is 2.47 W. The error margin in both cases was approximately 10%. Furthermore, in contrast to existing investigations, this experiment with the 8W@77K pulse tube cryocooler for recovering nitrogen is conducted under environmental pressure rather than overpressure, and it also takes into account the sensible heat issues caused by ambient pressure. The ratio of 1.81 W of sensible heat to 2.74 W of latent heat is 2:3, indicating that under ambient pressure conditions, the impact of sensible heat cannot be ignored. The key improvement methods are given to improve the success rate of the experiment. The research results provide a useful reference for the re-cycling and storage of liquid nitrogen.