The purpose of this work is to describe a method and apparatus that can be used to confirm the source strength of a large number of I‐125 seeds while maintaining sterility, accuracy, reproducibility, and time efficiency. Source strengths ranging from 0.395 to 0.504 U/seed were available for this study. Three different seed configurations were measured: loose, linked, and loaded needles. A third‐party 10% assay (NIST traceable) was provided. A custom stand was built out of aluminum to hold an exposure meter [Inovision (Fluke) 451P pressurized ion chamber] at 25 cm above the I‐125 sources to measure the exposure rate. The measurements were made in an operating room, and a sterile sheet was placed under the nonsterile aluminum stand on a sterile loading table. Seeds and needles were placed in a sterile tray for these measurements. Two hundred and six loose seeds in six batches (0.395, 0.395, 0.409, 0.444, 0.444, and 0.444 U/seed) and 1434 seeds in 10 batches containing various strands (0.444, 0.444, 0.444, 0.444, .0444, 0.466, 0.466, 0.504, and 0.504 U/seed) were measured. For the loose and stranded seeds, the average exposure rate per unit activity was measured to be 0.589 mR/h·U with a standard deviation of 0.017. Loaded needles were measured with an average exposure rate per unit activity to be 0.269 mR/h·U with a standard deviation of 0.014. We conclude that the method described here is capable of confirming a third‐party assay when performed on a large number of loose or stranded seeds in bulk. It is less reliable for preloaded needles.