Abstract The onsite calibration precision of the ocean thermometer can be efficiently raised by temperature scale transfer technology based on miniature fixed points. A new reproduction method of miniature Ga fixed points was studied for the sub-millikelvin precision measurement requirement. The onsite calibration system consists of a miniature Ga fixed point and a shipborne calibration device. The performance of miniature Ga fixed point is highly dependent on placement methods and the value-taking methods of phase transition temperature. This paper analyzed and compared the melting plateau of a miniature Ga fixed point with two placement methods and three value-taking methods of phase transition temperatures. The experimental results showed that utilizing horizontal placement and the average value method can achieve fixed point reproducibility within 1 mK. The performance of the miniature Ga fixed point was verified using the SBE 3plus thermometer from the SBE 911plus CTD system, with a thermometer accuracy of ±1 mK. The difference between the measured value by SBE 3plus and the Ga melting plateau value was within 0.5 mK. Thus, the study of the miniature Ga fixed point offers effective technical support for the onsite calibration of ocean thermometer.