At present, robust quality criteria and methods for the assessment of Peak inspiratory flow meter performance are lacking. A standard flow-volume simulator for quality control analyses of an inhalation assessment device was utilized with different simulated resistance levels in order to propose a quality testing method and associated standard for this device type. A standard flow-volume simulator was utilized to assess the performance of an In-Check DIAL® (Device I) and an intelligent inhalation assessment device (Device P) at a fixed volume and flow rate. Indices used to evaluate these two instruments included repeatability, accuracy, linearity, and impedance. Both devices exhibited good repeatability (<± 3 L/min). The difference between test results and standard simulator values for Device P was less than ± 5 L/min at resistance level R1 but higher than ± 5 L/min at resistance levels R2-5, while Device I were greater than 5 L/min at all resistance levels. The relative error for Device P was <± 10% at resistance levels R1, R2, and R4, but > 10% at resistance levels R3 and R5. The relative error values for Device I at all five resistance levels were > 10%. Device P passed the linearity test at the R2 resistance level, while Device I partially passed the linearity test at all five resistance levels. Standard monitoring methods and standards provide a valuable approach to the more reliable clinical assessment and application of these instruments.