The aim of this study was to measure patient and staff doses simultaneously for somecomplex x-ray examinations. Measurements of dose–area product (DAP) and entrance skindose (ESD) were carried out in a sample of 107 adult patients who underwent differentx-ray examinations such as double contrast barium enema (DCBE), single contrast bariumenema (SCBE), barium swallow, endoscopic retrograde cholangiopancreatography(ERCP) and percutaneous transhepatic cholangiography (PTC), and variousorthopaedic surgical procedures. Dose measurements were made separately foreach projection, and DAP, thermoluminescent dosimetry (TLD), film dosimetryand tube output measurement techniques were used. Staff doses were measuredsimultaneously with patient doses for these examinations, with the exception ofbarium procedures. The measured mean DAP values were found to be 8.33, 90.24,79.96 Gy cm2 for barium swallow, SCBE and DCBE procedures with the fluoroscopy timesof 3.1, 4.43 and 5.86 min, respectively. The calculated mean DAP was26.33 Gy cm2 for diagnosticand 89.76 Gy cm2 therapeutic ERCP examinations with the average fluoroscopy times of1.9 and 5.06 min respectively. Similarly, the calculated mean DAP was97.53 Gy cm2 with a corresponding fluoroscopy time of 6.1 min for PTC studies. The calculatedmean entrance skin dose (ESD) was 172 mGy for the orthopaedic surgical studies.Maximum skin doses were measured as 324, 891, 1218, 750, 819 and 1397 mGy forbarium swallow, SCBE, DCBE, ERCP, PTC and orthopaedic surgical procedures,respectively. The high number of radiographs taken during barium enema examinations,and the high x-ray outputs of the fluoroscopic units used in ERCP, were themain reasons for high doses, and some corrective actions were immediately taken.