Abstract Background Qualitative assessment of IgG-specific oligoclonal banding (OCB) in paired serum and cerebrospinal fluid (CSF) samples is useful for evaluating inflammatory conditions of the central nervous system such as multiple sclerosis. Here we characterized the performance of an IgG-specific method using isoelectric focusing on agarose gel matrix for the detection of OCB in paired serum and CSF samples. Methods Specimens of 20 paired, simultaneously collected CSF and serum patient samples were tested on Sebia (Hydrasys 2 System) provided by the National Institutes of Health (NIH) Diagnostic Laboratories. We tested these paired samples using the Helena IgG IFE-20 kit (SPIFE® Touch) method and the results were read by two independent individuals blinded to each other’s assessments. The results were compared to the NIH results for agreement. Positive Percent Agreement (PPA) and Negative Percent Agreement (NPA) were calculated. Analytical sensitivity of the Helena assay was assessed via serial dilutions of a CSF OCB positive sample (total IgG 4.5 mg/dL). Results Figure 1 shows serum and CSF pairs for the control (A), two example positive OCB patients (B & D) and a negative OCB patient (C). The Helena assay was concordant with Sebia for 11/12 positive OCB patients (PPA 91.6%) and 8/8 negative subjects (NPA 100%). 20-fold was the highest dilution that remained positive. Conclusions OCB detection using paired serum and CSF samples and isoelectric focusing electrophoresis is considered the preferred method. The Helena IgG IFE-20 assay is such a method and showed high PPA and NPA concordance with the Sebia System. Thus, the Helena OCB assay can be considered an accurate and analytically sensitive method for assessing OCB in serum and CSF.