This study tests if the pressure variation in the HFNC-system may allow for monitoring of respiratory rate and the pressure difference during breathing may be a marker of respiratory effort. A HFNC system (Fisher & Paykel Optiflow Thrive 950) was modified by adding a GE Healthcare D-Lite spirometry sensor attached to a respiratory module and a pressure transducer. Participants were instructed to breathe regularly, quickly and slowly during 4 different conditions (HFNC flow 30l/min and 70l/min and with an open and closed mouth). Respiratory rate was counted based on pressure variation shown on the monitor graphs and compared with the count by observation of the participant. The pressure difference between inspiration and expiration was tested for correlation with the respiratory rate, as a surrogate marker for respiratory effort. Twenty five participants were included in this study. False detection of apnea in pressure-based measurements occurred in 10% and 11% of the measurements with open mouth position at 30l/min and 70l/min HFNC-flow, respectively, but not with a closed mouth. The 95% Limits of Agreement were - 1.85;1.91, -13.72;9,88, -2.25;2.47, -30.32;19.93 for the conditions of 30l/min -closed mouth, 30l/min - open mouth, 70l/min - closed mouth and 70l/min - open mouth, respectively. There was a correlation between pressure difference and respiratory effort, except for the condition of 30l/min with open mouth. The pressure variation in the HFNC system allows for respiratory rate and effort monitoring, but requires further development to increase precision. ClinicalTrials.gov (NCT05991843).