The neutron time-of-flight (nTOF) detector, which is based on bibenzyl, an organic crystal scintillator with low afterglow, is used to diagnose the fuel areal density of target capsules by measuring the down-scattered neutron ratio. This paper presents a method for calibrating the sensitivity of bibenzyl scintillation detectors to 1–5 MeV down-scattered neutrons. The distributions of detector sensitivity are analyzed using the Monte-Carlo method. The analysis shows that most light output of neutrons below 5 MeV comes from recoil protons. Therefore, the sensitivity of the back-scattered neutron can be calculated with the 2.45 MeV neutron sensitivity and the relative responses of the 1–5 MeV protons. These responses are calibrated using a deuterium-deuterium implosion neutron source at a 100-kJ laser facility and a monoenergetic proton beam at a Van de Graaf accelerator, respectively. The total uncertainty of the calibration for 1–5 MeV neutron sensitivity ranges from 5% to 9%. This calibration method provides accurate measurements for the broad-spectrum neutron sensitivity of scintillation detectors.