Diabetic patients are at higher risk of recurrent adverse events following percutaneous coronary intervention (PCI) than the nondiabetics. Despite the introduction of new generation drug-eluting stents, their efficacy in the diabetics is still limited. To evaluate the efficacy of the Abluminus DES+ biodegradable polymer sirolimus-eluting stent in reducing neointimal hyperplasia in diabetic patients, compared to a durable polymer everolimus-eluting stent (DP-EES). A total of 131 patients with diabetes and coronary artery disease were enrolled in six Italian centers and randomized in a 2:1 fashion to PCI with Abluminus DES+ or DP-EES: 85 were assigned to Abluminus DES+ and 46 to DP-EES. The primary endpoint was optimal coherence tomography (OCT)-derived neointimal volume at 9-12 months. Secondary endpoints included OCT-derived neointimal area, neointimal volume obstruction and adverse clinical events. The primary endpoint, neointimal volume, did not differ between Abluminus DES+ and DP-EES (29.11 ± 18.90 mm3 vs. 25.48 ± 17.04 mm3 , p = 0.40) at 9-12-month follow-up. This finding remained consistent after weighing for the sum of stents lengths (1.14 ± 0.68 mm3 vs. 0.99 ± 0.74 mm3 for Abluminus DES+ and DP-EES, respectively, p = 0.38). Similarly, other OCT-derived and clinical secondary endpoints did not significantly differ between the two groups. Rate of target lesion failure was high in both groups (21.2% for Abluminus DES+ and 19.6% for DP-EES). This preliminary study failed to demonstrate the superiority of the Abluminus DES+ over the DP-EES in diabetic patients in terms of neointimal proliferation.