The present study aimed to evaluate magnetic resonance (MR) thermometry using proton resonance frequency shift (PRFS) during laser-induced thermotherapy (LITT), and to compare the results of using different sequences at a field strength of 7-Tesla to identify the optimal for use in ablation so that the surrounding healthy tissues may be protected from damaging in real time. LITT was applied to agarose gel phantoms and ex-vivo porcine brains. We reconstructed both magnitude and phase images to perform MR thermometry based on PRFS methods. We tested four different sequences: a gradient-echo (GRE), a segmented gradient-echo echoplanar imaging (EPI-GRE), a fast-low angle shot (FLASH), and a true fast imaging with steady precession (TRUFI). Temperature was monitored and verified using a fiber-optic thermometry device. All sequences showed good linear correlations (R = 0.97-0.99) between the measured temperature and the calculated MR-thermometry measurements. The phantom/porcine brain experiments revealed the temperature precisions at 1.53/0.69°C (GRE), 0.61/0.43°C (EPI-GRE), 1.64/1.32°C (FLASH), and 0.58/1.52°C (TRUFI), respectively. Furthermore, we performed a Bland-Altman analysis and the temperature accuracies were found to be - 1.32/- 0.60°C (GRE), 0.42/- 0.33°C (EPI-GRE), - 1.28/- 0.98°C (FLASH), and 0.14/0.46°C (TRUFI) in the phantom/porcine brain experiments, respectively. Our experiments recommend that EPI-GRE sequence be the best of the all sequences for MR temperature imaging with PRFS in the LITT on 7T magnetic resonance imaging (MRI) systems because of its relatively higher precision and accuracy.