A trueness-based EQA/PT program for high density lipoprotein cholesterol (HDL-C) was initiated. We analyzed the 4year EQA/PT program to overview the measurement standardization for HDL-C in China. Two levels of freshly frozen, commutable serum external quality assessment/proficiency testing (EQA/PT) materials were prepared and determined by reference measurement procedure each year. The samples were delivered to clinical laboratories and measured 15 times in 3days. The precision [coefficient of variation (CV)], trueness (bias), and accuracy [total error (TE)] were calculated and used to evaluate measurement performance. The pass rates of individual laboratories and peer groups were analyzed using the acceptable performance from the National Cholesterol Education Program (NCEP) and biological variation as the evaluation criteria. More than 60% of laboratories use heterogeneous systems, and there was a decrease in the percentage from 2016 to 2019. About 95, 78, and 33% of laboratories met the minimum, desirable and optimum TE criteria derived from biological variation. The pass rates were 87.0% (84.7-88.8%), 58.7% (55.3-62.4%), and 97.3% (95.6-98.3%) that met the acceptable performance of TE, bias, and CV of NCEP. The homogeneous systems had higher pass rates of TE, bias, and CV than the heterogeneous groups in 2016, but they did not show apparent advantages in 2017-2019. The trueness-based EQA/PT program can be used to evaluate the accuracy, reproducibility, and trueness of results. For some IVD manufacturers and individual laboratories, accuracy, especially trueness, are still problems. Efforts should be made to improve the situation and achieve better HDL-C measurement standardization.