Implementation of orthogonal double resonance precursor and neutral loss scans on the Mini 12 miniature rectilinear ion trap mass spectrometer is described, and performance is compared to that of a commercial Thermo linear trap quadropole (LTQ) linear ion trap. The ac frequency scan version of the technique at constant rf voltage is used here because it is operationally much simpler to implement. Remarkably, the Mini 12 shows up to two orders of magnitude higher sensitivity compared to that of the LTQ. Resolution on the LTQ is better than unit at scan speeds of ~ 400Th/s, whereas peak widths on the Mini 12, on average, range from 0.5 to 2.0Th full width at half maximum and depend heavily on the precursor ion Mathieu q parameter as well as the pump down time that precedes the mass scan. Both sensitivity and resolution are maximized under higher pressure conditions (short pump down time) on the Mini 12. The effective mass range of the product ion ejection waveform was found to be 5.8Th on the Mini 12 in the precursor ion scan mode vs. that of 3.9Th on the LTQ. In the neutral loss scan mode, the product ion selectivity was between 8 and 11Th on the Mini 12 and between 7 and 8Th on the LTQ. The effects of nonlinear resonance lines on the Mini 12 were also explored. Graphical Abstract ᅟ.