HLA-B27 is a major histocompatibility complex (MHC) class I antigen which exhibits strong association (90%) with ankylosing spondylitis. HLA-B27 detection in patients by flow cytometry is a widely used clinical test, performed on many different flow cytometer models. We sought to develop and validate a test conversion protocol for the HLA-B27 test performed on the BD FACSCanto to BD's newer FACSLyric flow cytometers. The development and validation experiments were performed using anti-HLA-B27*FITC/CD3*PE antibody-stained whole blood patient specimens. The anti-HLA-B27*FITC logarithmic median fluorescence (LMF) results on the BD FACSCanto were converted to median fluorescence intensity (MFI) values on the BD FACSLyric. Clustering of the HLA-B27 positive and negative values, using a 3rd order polynomial equation, resulted in a conversion of the BD FACSCanto cutoff values, negative (<150 LMF) and positive (≥160 LMF), to negative (<4530 MFI) and positive (≥6950 MFI) on the BD FACSLyric. Accuracy was assessed by comparing the flow results obtained on the BD FACSCanto and BD FACSLyric to a molecular PCR based assay. Additional validation parameters (compensation verification, intra- and inter-assay precision, and instrument comparison) were performed per the recommendations outlined in the Clinical and Laboratory Standards Institute (CLSI) H62 guidelines for validation of flow cytometry assays.