Hemolysis is a major pre-analytical concern for many laboratory analytes; however, instruments utilized for whole blood chemistries and blood gas measurements lack the ability to detect and measure the degree of hemolysis. This study evaluated the effect of hemolysis on 13 routine whole blood and blood gas analytes and compared visual assessments of hemolysis to measured hemolysis (H-index). Remnant whole blood samples (n = 85) were split into 2 portions and aspirated through a syringe one or more times. To induce hemolysis, a needle was affixed to the end of the syringe to provide shear stress, and a mock procedure without syringe was used as a control. Samples were analyzed on a Radiometer ABL800 series instrument, centrifuged, and the H-index of the plasma portion was measured. Two medical technologists recorded a visual categorization of the specimens as slightly, moderately, or severely hemolyzed. Hemolysis had a modest effect on metabolites and most cooximetry components, with percent bias within ±5% at all levels of hemolysis. Methemoglobin exhibited a larger overall negative bias, up to 13.3%. The absolute pH bias was fairly consistent (within 0.1 pH units) across all levels of hemolysis. As expected, potassium displayed a significant positive bias with increasing hemolysis. Sodium and ionized calcium displayed overall linear trends with a significant negative bias. Hemolysis can falsely increase or decrease certain blood gas analytes and lead to misinterpretation of results. Therefore, hemolysis detection capabilities are crucial for mitigating this effect and ensuring accurate results for patient care.
Read full abstract