We present the first results of a custom low-noise silicon photodiode detector designed for use with X-ray capillary optics. This system is intended to perform as a smaller, lighter, and low-powered alternative to currently existing X-ray telescope technologies. The detector has an active area with a diameter of 100 μm, and its small size leads to a low capacitance of well under 1 pF and a leakage current of 1 pA or less at the depletion voltage. We tested the detector with both a discrete preamp and an ultralow-noise front-end application-specific integrated circuit (ASIC). The front-end ASIC, developed at Brookhaven National Laboratory, was designed for low-capacitance detectors but had not been previously tested with a detector. Here, we discuss the detector design, the experimental setup, and the resulting detector performance. Using radioactive laboratory sources, we measure a full-width at half-maximum (FWHM) of 211 ± 8 eV at 5.9 keV with the ASIC and are able to set the threshold as low as 0.6 keV.