Intraosseous (IO) infusion, the pressurized injection of fluids into bone through a catheter, is a life-preserving resuscitative technique for treating trauma patients with severe hemorrhage. However, little is known regarding the application times, placement accuracy, and end-user ratings of battery-powered and manual IO access devices. This study was specifically designed to fill these knowledge gaps on six FDA-approved IO access devices. Three experienced U.S. Navy Emergency Medicine residents each placed commercially available 15-gauge IO catheters in cadaveric swine (Sus scrofa) proximal humeri and sternums in a randomized prospective experimental design. Devices included the battery-powered EZ-IO Rapid Infuser and the manual Jamshidi IO, PerSys NIO, SAM Manual IO, Tactical Advanced Lifesaving IO Needle (TALON), and PYNG First Access for Shock and Trauma 1 (30 trials per device, 10 per user, 210 total trials). Application times, placement accuracy in medullary (zone 1) and trabecular (zone 2) bone while avoiding cortical (zone 3) bone, and eight subjective user ratings were analyzed using ANOVA and nonparametric statistics at P < .05. The EZ-IO demonstrated the fastest application times, high rates in avoiding zone 3, and the highest user ratings (P < .0001). The TALON conferred intermediate placement times, highest rates of avoiding zone 3, and second-highest user ratings. The SAM Manual IO and Jamshidi performed poorly, with mixed results for the PerSys NIO and PYNG First Access for Shock and Trauma 1. The battery-powered EZ-IO performed best and remains the IO access device of choice. The present findings suggest that the TALON should be considered as a manual backup to the EZ-IO.