To assess the impact of scatter radiation on quantitative performance of first and second-generation dual-layer spectral computed tomography (DLCT) systems. A phantom with two iodine inserts (1 and 2mg/mL) configured to intentionally introduce high scattering conditions was scanned with a first- and second-generation DLCT. Collimation widths (maximum of 4cm for first generation and 8cm for second generation) and radiation dose levels were varied. To evaluate the performance of both systems, the mean CT numbers of virtual monoenergetic images (MonoEs) at different energies were calculated and compared to expected values. MonoEs at 50 versus 150keV were plotted to assess material characterization of both DLCTs. Additionally, iodine concentrations were determined, plotted, and compared against expected values. For each experimental scenario, absolute errors were reported. An experimental setup, including a phantom design, was successfully implemented to simulate high scatter radiation imaging conditions. Both CT scanners illustrated high spectral accuracy for small collimation widths (1 and 2cm). With increased collimation (4cm), the second-generation DLCT outperformed the earlier DLCT system. Further, the spectral performance of the second-generation DLCT at an 8cm collimation width was comparable to a 4cm collimation on the first-generation DLCT. A comparison of the absolute errors between both systems at lower energy MonoEs illustrates that, for the same acquisition parameters, the second-generation DLCT generated results with decreased errors. Similarly, the maximum error in iodine quantification was less with second-generation DLCT (0.45 and 0.33mg/mL for the first and second-generation DLCT, respectively). The implementation of a two-dimensional anti-scatter grid in the second-generation DLCT improves the spectral quantification performance. In the clinical routine, this improvement may enable additional clinical benefits, for example, in lung imaging.