To investigate the feasibility and accuracy of iodine quantification using PCD-CT in standard-pitch and high-pitch scanning at different scan parameters in a phantom model. Four inserts with known iodine concentrations (2, 5, 10, and 15mg/mL) were placed in the removable CT phantom and scanned using high-pitch (3.2) and standard-pitch (0.8) modes on PCD-CT. Two tube voltages (120 and 140 kVp) and four radiation doses (1, 3, 5, and 10mGy) were alternated. Each scan setting was repeated three times. Mean iodine concentration for each insert across three consecutive slices was recorded. Percentage absolute bias (PAB) was assessed for iodine quantification. A total of 96 acquisitions were conducted. In small phantom, the average for PAB was 2.96% (range: 1.75% ~ 4.56%) and 1.67% (range: 1.00% ~ 3.42%) for high-pitch and standard-pitch acquisitions, respectively. In large phantom, it was 3.72% (range: 1.75% ~ 5.97%) and 2.94% (range: 1.75% ~ 4.70%). Linear regression analysis revealed that only phantom size significantly influenced (P < 0.001) the accuracy of iodine quantification. The high-pitch scan mode in PCD-CT can be used to quantify iodine density with similar accuracy compared with standard pitch.
Read full abstract