Microporous polyvinylpyrrolidone-graphene-chitosan (PVP-G-CS) adsorbent beads were fabricated using a novel technique. The acidic PVP-G-CS mixture was aerated with an air pump, causing bubbles to form in drops that solidified in an alkaline solution to form the adsorbent beads. The beads were applied to simultaneously extract four parabens in beverage samples for analysis by high performance liquid chromatography-diode array detection. The results exhibited linearity in the range of 5.0 – 1000 µg kg−1 under optimal conditions. The ranges of the limits of detection and limits of quantitation were 4.3 – 7.3 µg kg−1 and 14.3 – 24.3 µg kg−1, respectively. The developed beads demonstrated good repeatability (RSD < 5.7 %, n = 6) and were reusable for up to 15 times in juice samples (RSD ≤ 5.3, n = 3). Recoveries of the target parabens from beverage samples ranged from 72.2 % to 102.8 %. Two alcoholic beverages and four juice samples were found to contain paraben concentrations ranging from 7.2 to 14.2 µg kg−1.