The bipolar radiofrequency ablation(B-RFA) strategy was increasingly used to target deep intramural re-entrant foci responsible for the arrhythmia not ablated by conventional unipolar RFA / sequential unipolar RFA. Lesional characteristics of various bipolar configurations were largely unknown. To investigate the lesional geometry in relation to various factors to determine the most effective ablation strategy that minimises steam pops and achieves transmurality. To assess the temperatures at the return electrode. A custom-made validated ex-vivo bipolar ablation model was used to assess lesion formation. The myocardial sample was placed between two ablation catheters in four different orientations. Lesions were created using different power (30W, 40W, 50W) and time settings(30, 40 and 50s) with different catheter orientations. Data was analysed using binary logistic regression and multiple linear regression. Among 107 lesions, The volume of the active catheter lesion (266 +/- 137mm^3) significantly differed from their return electrode counterparts (130 +/- 91.8mm^3) (p < 0.001), and the temperatures at the return electrode end were lower than at the active electrode (p = 0.004). Higher power and longer duration application led to more frequent steam pops (p < 0.001), while true parallel configuration resulted in fewer steam pops (p < 0.001). A custom model without ground electrode temperature monitoring is safe and cost-effective. The safest strategy is a true parallel configuration with an inter-electrode distance of at least 15mm and a power of 30W to 40W, which generates lower steam pops and better transmurality.