Beam-matched linear accelerators (LA's) require accurate and precise dosimetry for fractionated stereotactic treatment. In this study, the beam data were validated by comparing the three-beam-matched LA's measured data and the vendor reference data. Upon its validation, the accuracy of the volumetric dose delivery for eighty patient-specific fractionated stereotactic treatment plans was evaluated. Measurements of the percentage depth dose (PDD), beam profiles, output factors (OFs), absolute output, and dynamic multi-leaf collimator (MLC) transmission factors for 6MV and 10MV flattening filter (FF) and flattening filter-free (FFF) photon beams were obtained from three-beam-matched LA's. The patient-specific quality assurance evaluation for all eighty plans was performed using PTW Octavius 1000 SRS™ array detectors for two-dimensional (2D) fluence measurement. The following 2D gamma passing criteria were used: 1%/1mm, 2%/1mm, 1%/2mm, 2%/2mm and 3%/2mm. In all three LA's, gamma analysis for PDD and profile were above 97% with gamma criteria of 1%/1mm. The differences OFs, absolute output, and dynamic MLC transmission factors were less than ± 1% of base value. For all eighty cases, the median passing rates on the three LA's were above 76%, 88%, 92%, 96%, and 98% for the above-mentioned gamma criteria of the three LA's. The beam-matched LA's showed good agreement between the measured and treatment planning system (TPS) calculated values for fractionated stereotactic VMAT plans with 6MV and 10MV (FF and FFF) photon beams. Patients can be shifted and treated on any beam-matched linac without the need of re-planning.