Background: Cancer, defined as the uncontrolled proliferation of malignant cells, is a serious hazard. Radiotherapy, which employs high-energy radiation, seeks to break cancer cell DNA. Due to the significant consequences involved, accuracy is of utmost importance. Before commencing radiation therapy, thorough quality assurance procedures are carried out. The Percentage Depth Dose (PDD) is employed to ensure accurate dose measurement. The research utilized Cobalt-60 Teletherapy, which produces gamma radiation at energy levels of 1.17 MeV and 1.33 MeV, to assess the various field widths observed in PDD. In this experiment, the percentage depth dose (PDD) was evaluated by using a Farmer chamber, water phantom, PC electrometer, ionization chamber, computer software and barometer. The chamber was used throughout a range of field sizes, which varied from 5x5cm2 to 20x20cm2.The TRS-398 approach was employed to assess the radiation dose. Comparable findings were observed when conducting comparisons with published PDD data for cobalt-60 beams. The accumulation dose zone was found to be located approximately 0.5 cm below the surface, coinciding with the maximum (Zmax) dose site. The absorbed amount of radiation resulted in a decrease in the percentage depth dose (PDD) below the given depth. DOI: https://doi.org/10.52783/pst.643