Biological valves are becoming more frequently used in aortic valve replacement. While several reports have evaluated the performance of biological valves, echocardiography studies during exercise stress remain scarce. Furthermore, no current reports compare rate changes in the aortic valve area of biological valves under increased exercise load. Here, we performed exercise stress echocardiography in patients after AVR with Trifecta or Inspiris valves and compared the rates of change in aortic valve areas (AVA). In addition, hydrodynamic analysis at rest was conducted with four-dimensional flow magnetic resonance imaging (4D-flow MRI). Exercise stress echocardiography was performed in seven Trifecta and seven Inspiris patients who underwent AVR at our hospital while 4D flow MRI was performed in all but two Trifecta cases. Comparing the percentage change in AVA when loaded to 25W versus at rest, Trifecta was greater than Inspiris (28.7 ± 36.0 vs - 0.8 ± 12.4%). The smaller AVA at rest was considered causative for this. Meanwhile, Trifecta systolic energy loss in the prosthetic valve segment on 4D-flow MRI (97.5 ± 35.9 vs 52.7 ± 25.3mW) was higher than Inspiris. The opening of the Trifecta valve was considered to be restricted at rest and this may reflect the current reports of early valve degradation requiring reoperation. Taken together, we observed that the Trifecta design may promote faster wear due to higher valve stress.