A streamlined, precise, and dependable method for the concurrent quantification of bromate has been established, employing ion chromatography in conjunction with electrospray ionization tandem mass spectrometry (IC-ESI-MS/MS). In an effort to expedite sample analysis duration, the AG18 short column was utilized for rapid separation of the specified analytes. Utilizing a blend of water and acetonitrile as the mobile phase within an optimized gradient elution setting, it was possible to achieve satisfactory resolution in under 3 minutes. Specifically, the mixing of acetonitrile in water samples improves the sensitivity and immunity of the method. This phenomenon may be attributed to the similar proportions of acetonitrile in the sample and the mobile phase, which effectively reduces the viscosity and surface tension of the droplets during the ESI process. Detection limits and quantification thresholds were 0.14 µg/L and 0.41 µg/L, respectively. Analyte recoveries, when spiked into tap water between 2.5 µg/L to 50 µg/L, ranged from 69.5 % to 116.7 %; precision metrics, measured as intra- and inter-day relative standard deviations, spanned 5.1–7.9 %. The devised approach encountered minimal interference from prevalent water matrix constituents, including chloride (50–200 mg/L), sulfate (25–200 mg/L), bicarbonate/carbonate (1–6 mM), and natural organic matter (1–8 mg/L). In environmental contexts, this methodology effectively tracked the production of bromate in controlled ozonation experiments conducted in the laboratory. Segmented ozone dosing reduced bromate production when the solution contained low concentrations of DOC, which provides an alternative strategy for controlling bromate production.