Routine defibrillation threshold testing (DFT) of transvenous implantable defibrillators (ICDs) has largely been in decline. In patients with non-transvenous ICDs that utilize subcutaneous and pleural ICD leads, serial DFT testing can detect a significant number of failures. Data about the utility of follow-up defibrillation safety margin testing (DSM) testing in pediatric patients and young adults with an epicardial ICD are lacking. Patients aged<25 years old who underwent epicardial ICD placement at Mayo Clinic from 2014 to 2023 with at least one follow-up DSM test were included. The patients were divided into a "routine" (R) and "clinically indicated" (CI) group based on the index of clinical concern. Inadequate DSM was defined as unsuccessful defibrillation at an output of less than 10 J below the maximum output of the device. The purpose of this study was to assess the utility of follow-up DSM testing. An epicardial ICD system was placed in 122 patients. A total of 26 patients met inclusion criteria and underwent a total of 47 DSM follow up tests. Inadequate DSM occurred in 1/33 (3%) in the R group and 2/14 (14%) DSM tests in the CI group. The median follow-up period was 54 and 36 months for the R and CI group, respectively. Our data suggest that epicardial ICDs are reliable and routine follow-up DSM testing may not be necessary for all patients. DSM testing should be performed in individuals with epicardial ICD systems when there is clinical concern about lead or coil performance.