BackgroundLeft Bundle Branch Area Pacing (LBBaP) is a cardiac pacing technique designed to mimic the natural conduction system of the heart. Traditional right ventricular apical pacing has been associated with increased risks of heart failure and atrial fibrillation. This study investigates the stability and safety of LBBaP using the Selectra 3D sheath (Biotronik) with an inclined angle for implanting the Solia S lead (Biotronik, SE & Co, KG).MethodsA single-center retrospective study was conducted on 25 patients who underwent LBBaP implantation using the Selectra 3D sheath at our hospital. The procedure involved inserting the Solia S lead into the interventricular septum at an inclined angle. Surgical and postoperative data were collected, including the success rate, depth and angle of electrode insertion, complications, and follow-up data.ResultsThe success rate of LBBaP implantation was 92%. The length of electrode insertion into the interventricular septum ranged from 12 to 23.0 mm, with an average of 18.1 ± 3.08 mm. The angle formed between the electrode and the septum ranged from 0° to 57.3°, with an average of 35.14°±14.31°. During the 3-month follow-up period, pacing parameters remained stable, and no complications were reported.ConclusionsLBBaP implantation using the Selectra 3D sheath with an inclined angle for the Solia S implantation demonstrates stability and safety. The procedure boasts a high success rate and offers an effective option for LBBaP implantation.