Abstract Background Photon-counting detector computed tomography (PCD-CT), which allows exclusion of electronic noise, shows promise for significant dose reduction in coronary computed tomography angiography (CCTA). This study aimed to assess the radiation dose and image quality of CCTA using PCD-CT, combined with high-pitch helical scanning and an ultra-low tube potential of 70 kVp and investigate the effect of a sharp kernel on image quality and stenosis assessment in such an ultra-low dose CCTA setting. Methods Forty patients (65% male) with stable heart rates and no prior coronary interventions were included. Data on CT dose index volume (CTDIvol) and dose length product (DLP) were collected, with effective radiation dose estimated using a conversion factor of 0.014. Images were reconstructed using kernels of Bv64 and Bv40 for image quality and stenosis assessment. Results The mean CTDIvol, DLP, and effective dose of CCTA were 1.72±0.38 mGy, 29.1±6.8 mGy·cm, and 0.41±0.09 mSv, respectively. Image quality was similar (p=0.75) between the two kernels, with over 95% of segments achieving a rating of good image quality for both kernels. Additionally, 42% of non-calcified plaques showed an increased stenosis severity from Bv40 to Bv64 (p<0.001), while 25% of calcified plaques exhibited a decreased severity (p<0.001). Conclusion PCD-CT technology with high-pitch helical scanning and the tube potential of 70kVp can provide CCTA with ultra-low radiation exposure (DLP, 29 mGy·cm). The noise reduction capability of PCD-CT allows the use of a sharp kernel even in this low-dose CCTA setting without compromising image quality, potentially improving the evaluation of coronary artery stenosis.