In this work, we synthesized and characterized the Methylhydrazinium Lead Trichloride MHyPbCl3(CH3NH2NH2PbCl3) perovskite as a fast neutron detector. The high hydrogen density of MHyPbCl3 enables efficient energy conversion from a fast neutron into a recoiled proton through the 1H(n,n)1H elastic scattering interaction, thereby, allowing for direct charge detection. Through IV characterization and X-Ray excitation, the crystal demonstrated a high resistivity at 4.43E11Ωcm and a good mobility-lifetime product (μτ) of 9.1E-3 cm2V, respectively, under C60/BCP/Cu and Au contact configuration to form an ohmic-ohmic detector. The crystal showed a good sensitivity to X-rays using an x-ray tube. The feasibility of the direct neutron conversion detector is demonstrated using the fast neutron beam at a Research Reactor. Waveforms from a charge-sensitive pre-amplifier showed distinct radiation-induced pulses from the MHyPbCl3 detector in response to the reactor neutron beam. Using a thermal neutron filter and gamma shielding in the beam, we showed that the pulses produced were more likely from neutron interactions, despite the Pb containing MHyPbCl3 is also sensitive to gamma-rays. With those fast neutron pulses, a post-pulse processing code was used to conduct pulse height analysis (PHA) and reconstruct an energy spectrum.