We present an intravascular ultrasound (IVUS) transducer array designed to enable shear wave elasticity imaging (SWEI) of arteries for the detection and characterization of atherosclerotic soft plaques. Using a custom dicing fixture, we have fabricated single-element and axially-segmented array transducer prototypes from 4.6-Fr to 7.6-Fr piezoceramic tubes, respectively. Focused excitation of the array prototype at 4 MHz yielded a focal gain of 5× in intensity, for an estimated 60 mW/cm2 [Formula: see text] and 1.6-MPa negative peak pressure at 4.5-mm range in water. The single-element transducer generated a peak radial displacement of [Formula: see text] in a uniform elasticity phantom, with axial shear waves detectable by an external linear array probe up to 5 mm away from the excitation plane. In a vessel phantom with a soft inclusion, the array prototype generated peak displacements of 2.2 and [Formula: see text] in the soft inclusion and vessel wall regions, respectively. A SWEI image of the vessel phantom was reconstructed, with measured shear wave speed (SWS) of 1.66 ± 0.91 m/s and 0.97 ± 0.59 m/s for the soft inclusion and vessel wall regions, respectively. The array prototype was also used to obtain a SWEI image of an ex vivo porcine artery, with a mean SWS of 3.97 ± 1.12 m/s. These results suggest that a cylindrical intravascular ultrasound (IVUS) transducer array could be made capable of SWEI for atherosclerotic plaque detection in coronary arteries.