Continuous deep septal pacing and signal recording during implantation of left bundle branch pacing (LBBP) lead enables to monitor beat-to-beat changes of electrocardiogram (ECG) and myocardial current of injury (COI) as the lead crosses the septum. This study aimed to characterize patterns of continuous QRS, ST-T, and COI change for monitoring of the lead depth and instantaneous determination of the obtained capture type (LBBP vs left ventricular septal pacing [LVSP]). The ECG and COI during lead implantation were scrutinized for sudden changes of V6 R-wave peak time, V1initial and terminal R-wave amplitude, V3-V6 R-wave amplitude, repolarization pattern and S-wave amplitude in I, V5-V6, and COI drop. The sudden and gradual transition patterns were diagnosed depending on the presence or absence of the above beat-to-beat ECG phenomena, respectively. A total of 212 pacemaker recipients were analyzed; LBBP and LVSP were obtained in 77.4% and 22.6%, respectively. There were 4.7 ± 2.1 and 0.2 ± 0.6 beat-to-beat phenomena in LBBP and LVSP patients, respectively. The sudden transition pattern, recognized in 80.7%, had sensitivity and specificity for LBBP diagnosis of 98.8% and 81.2%, respectively. A sudden drop of COI (29.4 ± 8.5mV to 12.8 ± 4.9mV) was observed in 53.9% patients (LBBP was simultaneously obtained in 92.7%). Capture of left bundle branch during lead penetration is a beat-to-beat phenomenon. Two transseptal transition patterns were identified: 1) sudden, which is typical for obtaining LBBP; and 2) gradual, which is typical for obtaining LVSP. A sudden COI drop, a very observable phenomenon, also identified reaching the left subendocardial area.