MAGnetic Expansion Control (MAGEC) rods can prevent repeated lengthening operations for scoliosis patients. However, there have been several Field Safety Notices issued, including a worldwide product recall due to actuator endcap separation. We aimed to review adverse events reported to the Food and Drug Administration (FDA) regarding MAGEC rods, focusing on MAGEC X. Reports submitted to the Manufacturer and User Facility Device Experience database in relation to MAGEC devices were accessed and analysed using R Statistical Software. Exclusion criteria included duplicate and literature review reports (n = 54). Free-text data were analysed using inductive content analysis. 1016 adverse events were reported to 11/30/2023. 99.0% (1006) were submitted by the manufacturer. Reports primarily arose from the UK (465, 45.8%) or US (421, 41.4%). From free-text data the most frequent adverse events were distraction mechanism failure (573), device wear (272), and actuator seal damage (180). Rod fracture (n = 48) was not significantly associated with rod diameter (≤ 5.0mm or > 5.0mm), p = 0.736. 234 reports referenced MAGEC X devices; actuator endcap separation was identified in 41.9% (99). Other events include failure of distraction (63), surface damage (31), and rod fracture (15). On 06/30/2020 MAGEC X2 received FDA approval. Twenty reports reference devices manufactured after this date, seven describe distraction mechanism failure; notably there are no reports of endcap separation. These data represent the largest series of adverse events reported for MAGEC rods, including significant new data regarding MAGEC X. As well as endcap separation, failure of distraction, surface damage, and rod fracture were reported.