Abstract Small linear ion trap mass spectrometers are often combined with direct ionization mass spectrometry (DIMS) techniques for detection and analysis. However, DIMS is susceptible to the working environment and matrix effects, resulting in reduced sensitivity, which places high demands on the performance of tandem mass spectrometry of small linear ion trap mass spectrometers. Therefore, the effects of vacuum pressure, collision q-value, collision voltage, and collision time on the dissociation efficiency of the ion trap are experimentally explored in this paper to enhance the performance of tandem mass spectrometry for small linear ion trap mass spectrometers. The findings indicate that as the vacuum pressure increases, dissociation efficiency first increases rapidly and then remains stable, and the peak dissociation efficiency of 56% is attained at a vacuum pressure of 0.31 mTorr. Increasing the collision q value, dissociation efficiency first increased slowly, then increased rapidly, and finally remained almost unchanged. When the collision q value was increased from 0.26 to 0.35, dissociation efficiency increased from 9% to 56%. Meanwhile, the maximum collision voltage corresponding to the optimal dissociation efficiency gradually decreases as the collision q value increases. With the increase in collision time, dissociation efficiency increases rapidly and then remains stable. Therefore, studying the factors affecting dissociation efficiency points the way to enhancing the efficacy of tandem mass spectrometry with small linear ion trap mass spectrometers.