For some cochlear implants (CIs), it is possible to focus electrical stimulation by partially returning current from the active electrode to nearby, intra-cochlear electrodes (partial tripolar (pTP) stimulation). Another method achieves the opposite: "blurring" by stimulating multiple electrodes simultaneously. The Panoramic ECAP (PECAP) method provides a platform to investigate their effects in detail by measuring electrically evoked compound action potentials and estimating current spread and neural responsiveness along the length of the CI electrode array. We investigate how sharpening and broadening the electrical current spread are reflected in PECAP estimates. PECAP measurements were recorded at most comfortable level in 12 ears of Advanced Bionics CI users. Focused thresholds were also determined. For the electrodes with the highest and lowest focused thresholds, additional PECAP measurements were recorded while stimulating in pTP mode and in "blurred" mode with 3 or 5 adjacent electrodes simultaneously stimulated. Current spread and neural responsiveness were then estimated along the electrode array using PECAP. PECAP revealed increased current spread estimates across participants for blurred stimulation of the targeted electrodes towards the apex of the cochlea. Variable results for pTP stimulation were found, with two of eight ears appearing to drive a small group-level effect of increased current spread. When stimulating multiple electrodes simultaneously, PECAP detected localized increases in current spread towards the apex (but not the base) of the cochlea. pTP stimulation showed mixed effects on PECAP current spread estimates. These findings are in line with behavioral speech perception studies and have implications for cochlear implant optimization.