Quantitative protein mass spectrometry (MS) is ideally suited for precision diagnostics and for reference standardization of protein analytes. At the Leiden Apolipoprotein Reference Laboratory we apply MS strategies to obtain detailed insight into the protein-to-peptide conversion in order to verify that quantifier peptides are not partly concealed in miscleaved protein backbone. Apolipoprotein(a) (apo(a)) was digested in a non-optimal manner to enhance the number of miscleaved peptides that were identified by high resolution liquid chromatography tandem-MS measurements. The protein-to-peptide conversion was carefully mapped with specific attention for miscleaved peptides that contain an apo(a) quantifier peptide. Four different isotopologues of each apo(a)-quantifier peptide were applied to evaluate linearity of internal peptide standards during measurement of specific real-life samples. Two apo(a) quantifier peptides that were concealed in two different miscleaved peptides were included into a multiple reaction monitoring list in our targeted MS-based apo(a) quantifications to alert for potential protein digestion discrepancies. The presence of miscleaved peptides could be ruled out when applying our candidate reference measurement procedure (RMP) for apo(a) quantification. These data further corroborate the validity of our apo(a) candidate RMP as higher order method for certification of commercial Lp(a) tests that is endorsed by theInternational Federation of Clinical Chemistry and Laboratory Medicine. MS-based molecular detection and quantification of heterogeneous apo(a) proteoforms will allow manufacturers' transitioning from confounded lipoprotein(a) [Lp(a)] mass levels into accurate molar apo(a) levels.