In this study, we report our physics commissioning experience for clinically implementing a GRID collimator based Spatially Fractionated Radiation Therapy (SFRT). For SFRT commissioning, we utilized a commercially available brass GRID collimator (DotDecimal, Sanford, FL) that is mounted on a blocking tray for positioning into the accessory mount of the Linac. The brass GRID is 7.62 cm thick and weighs 15.8 kg. The GRID has a total of 149 divergent holes arranged in a hexagonal pattern (1.43 cm in diameter and hole-centers spaced at 2.11 cm when projected at the isocenter distance). The GRID encompasses a maximum field size of 25 × 25 cm at the isocenter. A commercial 48 × 48 × 48 cm water phantom scanning system was used to collect GRID output factors (GRID field to open field ratio), depth dose and beam profile data. Output measurements were performed using a 0.13 cm3 active volume ion-chamber and beam scans were obtained with a diode detector. Data was collected for both flattening and flattening-free beams of nominal energies 6 MV and 10 MV photons. The measurement depths were at dmax (1.5 cm for 6 MV and 2.5 cm for 10 MV), 5-cm and 10-cm respectively. For each energy and depth of measurement, collimator settings were varied from 5 × 5 cm to 28 × 28 cm. From scan profiles at different depths, the valley (lowest) to peak (highest) dose ratios (VPDR) were calculated. A commercial treatment planning system (TPS) was used to test the accuracy of dose calculations with GRID. This was accomplished by importing vendor generated DICOM RT file into the TPS. A block transmission factor of 7% for 6 MV and 10.2% for 10 MV energy beams were applied. All measured data were compared with corresponding TPS calculated data. Test patient treatment plans with GRID were created in TPS and planned distributions were verified using a commercial detector array with 2.5 mm detector spacing. The VPDR, expressed as %, are presented in Table 1. Measured and TPS calculated output factors agreed within 2% for 6 MV and within 3% for 10 MV photon beams. Percent depth doses were lower in magnitude for GRID field compared to open field for all energies studied (for e.g., 6 MV depth dose at 10 cm depth for 10 cm x10cm field 62% for GRID field vs. 66% for open field). Measured and calculated GRID beam profiles agreed within 5% dose difference and 1 mm distance-to agreement. For all test cases, the planned vs. measured dose distributions passed at an average gamma passing rate of 96.5% using a 3% dose difference/ 3 mm distance to agreement and 10% threshold criteria. The Dot Decimal GRID collimator provides a simple way of achieving SFRT in the clinic, albeit heavy to use and has an irradiation field size limitation of 25 cm × 25 cm.
Read full abstract